Eppendorf 5415D Error 3

Eppendorf 5415D Error 3.   Does anybody know of a vendor for Eppendorf parts other than the OEM?

avatar placemark

Asked by

Jkirk
Report this Post
avatar placemarkReply by
Green BioMed LLC

Hi,

This error is due to speed sensor error or the motor inside. we fix error 3 on eppendorf 5415d. we also have a service of rebuilding the centrifuge motor. sometime you only need to change one part. Please let us know if we can help

Laboratory Equipment Service & Repair - Green BioMed LLC

Thanks,

Green BioMed

sales@greenbiomed.com

Report this Post
avatar placemarkReply by
tub57042_2227

Error 3 is related to the speed sensor or motor as others have already stated.  Sometimes this error can be something simple.  For example, as the motor shock mounts wear over time (or if the end user doesn't balance the rotor properly) the motor takes on a lot of vibration.  Sometimes its bad enough to shake the motor connector loose.  I just fixed one like this recently.  So always check the easier issues first like the motor connector and the speed sensor connector to see if they are seated properly.  