7694 HS with 6890 GC/5973N MS

I am trying to connect an Agilent 7694 Headspace sampler to a 6890GC/5973N MS system. The gas line connection seems pretty straightforward but I am not sure about what types of cables/connections need to go from the Headspace to the GC. I have been trying to decipher the manuals but they are not as helpful as I had hoped. Can anyone help ?

8 months agoReply from Everhart
avatar placemark I agree toDan Ste.Marie.. You need that thing to connect that!our site

9 months agoReply from Dan Ste.Marie
RIChemistYou need an Agilent G1290-60570 remote cable. Dan

Agilent Technologies - 7694 headspace Sampler
by Agilent Technologies

Agilent Technologies - 7694 headspace Sampler
DESCRIPTION
7694 Headspace Sampler

7694 Headspace Sampler