V20S volumetric titrator use for sampling unknown water content

My sample can run anywhere between 10 ppm and 180 ppm. Can the V20s titrator give repeatable results in these ranges. Which Titrant is recommended? What settings would be recommended, sample size, control band, ipol, etc?

13 days agoReply from jburkman11
LabCat You will needs about 10-20 grams sample if using the 1-comp5 titrant

13 days agoReply from LabCat
avatar placemarkUsing a 1-comp 5 titrant.Had on hand). Testing refrigerant gas.
13 days agoReply from jburkman11
LabCat What type of sample are you testing? Does it contain any ketones?

Keep all method parameter the default. However would recommend using a 1-comp2 titrant, would still require about 2 gram sample size for accurate results.