V20S volumetric titrator use for sampling unknown water content

My sample can run anywhere between 10 ppm and 180 ppm. Can the V20s titrator give repeatable results in these ranges.   Which Titrant is recommended?  What settings would be recommended, sample size, control band, ipol, etc?

avatar placemark

Asked by

LabCat
Report this Post
jburkman11Reply by
jburkman11

What type of sample are you testing? Does it contain any ketones?

Keep all method parameter the default. However would recommend using a 1-comp2 titrant, would still require about 2 gram sample size for accurate results.


Jennifer Burkman
METTLER TOLEDO
www.mt.com  
Report this Post
avatar placemarkReply by
LabCat
Using a 1-comp 5 titrant. Had on hand). Testing refrigerant gas.
Report this Post
jburkman11Reply by
jburkman11

You will needs about 10-20 grams sample if using the 1-comp5 titrant


Jennifer Burkman
METTLER TOLEDO
www.mt.com  