Energy Scan Failure, Instrument Error, Establish Scan Time Out, Fatal Scan Error

Would you please adivse on the nominal energy?

I replaced the source in 2013 and laser in 2014. Occasionally, our Spectrum 100 will display an error message: "Instrument Validation: Energy Scan Failure, Instrument Error, Establish Scan Time Out, Fatal Scan Error." What can cause this error message?

What is the specification for the energy output? The Energy display shows 593 and the Centre Burst min-max is 0.13.

17 days agoReply from MicheleS
avatar placemark Thank you for responding to my question. An UATR (diamond/ZnSe) is installed.

17 days agoReply from Berridit
MicheleSIt's unlikely to be a source problem. The expected, average life of a laser is 4 years. If this energy you report is with an open beam, then you have a more complex issue. The energy I would expect with and open beam, with any accessories removed, would be >5500 counts. I depends on the type of accessory plugged in that will determine the energy otherwise. Scan errors can be caused by weak lasers. You can email me if you need further assistance. Iberridge@isocount.com

PerkinElmer - Spectrum 100 FTIR Spectrometer
by PerkinElmer

PerkinElmer - Spectrum 100 FTIR Spectrometer
DESCRIPTION
Enables much faster measurement of even the most demanding samples such as highly absorbing or poorly reflecting material.

The PerkinElmer Spectrum 100 Series FT-IR spectrometers are the gold standard in IR spectroscopy for materials testing, academia and research applications. The new Spectrum 100S version demonstrates the highest sensitivity in its class enabling much faster measurement of even the most demanding samples such as highly absorbing or poorly reflecting materials. While for the more cost-conscious laboratory , the Spectrum 100R combines ease of use, reliability and performance at a price normally seen among instruments with a fraction of the 100R’s analytical power. Sampling speed and system stability, backed by PerkinElmer's knowledge, experience and responsive service and support, enable laboratories to achieve the highest quality and reproducible results with ease. Nowhere else will you receive the same confidence in an infrared system that you gain with the PerkinElmer Spectrum 100 FT-IR systems.

Advertisement