Discussion Category: UV/VIS Spectrophotometers
power fail
why I have this problem
Asked by
mjimenez
Reply by
CHANDRA
Hi MJIMENEZ
pl refer the below:
Power fail reported on the Agilent ChemStation If your Agilent ChemStation can not connect to the spectrophotometer using LAN communication, please perform the following troubleshooting steps. Verify that basic communication is possible Use the ping command, to verify that the IP address is operational. 1 From the Start menu in the Task Bar, select Start > Programs > MS-DOS Command Prompt 2 Type ping <IP-number> and press Enter to verify communication of the JetDirect Card and the LAN adapter in your PC. <IP-number> needs to be replaced by the appropriate IP address (e.g. 10.10.10.101) or host name. 3 The command ping will send a request for reply to the IP address, bypassing part of the Windows TCP/IP settings. A successful ping would look like this: Reply from 10.10.10.101: bytes=32 time<10ms TTL=128 If the message request timed out is displayed, the IP address can not be reached by the ping command. 4 If the ping request was answered succe ssfully by the instrument, you need to verify that your Windows TCP/IP settings are correct for the selected network, especially the subnet mask and gateway settings should be checked. Identify by host name If you identify the Agilent 8453 system by host name, ensure that the used host name and IP address are correctly setup in the DNS server. Try using the IP address in the Configuration EditorReply by
Danko
I tired all of those state above but the problem is still occurred.
Add your reply: