Radiometer D53
We have a Beckman XL-100K that is giving a D53 error. The temperature reading is fine when the vacuum >100u, but is erratic when vacuum is
We suspect the radiometer 'collector' or the Peltier unit under (and part of the radiometer assembly) is damaged, although we see no indications even under a microscope.
We have tried changing vacuum pump and diffusion pump oils, cleaning any collected oils under the chamber can, removing the radiometer and cleaning it (and the windows film) with 95% ethanol (being sure to gently clean under the BB-wafer keeping the wires intact) and allowing the radiometer to dry, and then re-seating the radiometer (and of course the can) with heat-sink paste. However, we still see the same issue.
We are hesitant to replace it straight off - a new radiometer is very expensive!
Since the radiometer is reading 'something' (as erratic as the readings are) is it more likely a board, filtering cap, transistor/ic/regulator, stray field or other issue? Can anyone point us to a particular component (other than the radiometer) that is a common cause of the symptoms?
Does anyone have a service manual and/or electrical schematic?
Thank you in advance!