Baseline troubleshooting

Hello,

 

 

we are currently using skalar at a remote location, and about two days ago, we are having trouble with the baseline. TCN always stay about 1.2-1.3 and WAD 1.6-1.7. But now TCN is sitting at 2and WAD is at 3. The baseline is also unstable, with blip like picks all over, thats before using any in house prep solutions. only using DI water. what will be the possible issue here?, we are pretty sure that all solutions lines are flowing well and the glass coils are also clean. We have limited information on possible issues to encounter with the instrument. Can you please help us trouble shoot any possible cause that can affect the baseline and the instrument performance.

2 years agoReply from Ridha
avatar placemarkle manque de mouillant dans les réactifs, vérifier la date limite d'utilisation, l'usure des tubes de pompe, saleté et usures des galets de la pompe, les positions des tubes de pompe,veillissement du détecteur,
2 years agoReply from Shane13
avatar placemark Sometimes when a baseline is unstable it can be an electrical issue. It could be as simple as replacing the light bulb(which are unfortunately expensive)

Skalar Analytical - SAN++
by Skalar Analytical

Skalar Analytical - SAN++
DESCRIPTION
The San++ unit is a automated wet chemistry analyzer featuring over 300 approved methods for water, beer, wine, malt and industrial samples.

The San++ Automated Wet Chemistry Analyzer also known as Continuous Flow Analyzer has been designed as a modular system to meet the needs of laboratories which process a few samples or large volumes of samples. Based on the Continuous Flow Analysis (CFA) or also called Segemented Flow Analysis (SFA) technique, up to 16 analytical measurements can be made on a single sample simultaneously. With the addition of rinsing valves and software the San++ system can automate; start-up, shut-down, dilutions, reruns, clean-up, and raw data storage.          Skalar, the world leader in wet chemistry technology offers, over 300 proven applications utilizing various combinations of automatic dilutions, additions, mixing, heating, dialysis, extractions, distillation, digestion, phase-separation, hydrolysis, ion-exchange/reduction and more. Skalar has developed a proven systematic approach to providing a quality system starting with complete applications engineered on customer samples to provide the best possible configuration to the customer needs and requirements, training and installation by qualified specialists and complete application and service support.

Advertisement