Discussion Category: UV/VIS Spectrophotometers
Beckman DU650 Sudden change in photometric accuracy
I had a problem with wavelength cal on a Beckman DU650.
I aligned the internal slit and that problem went away.
In checking calibration I found that the photometric accuracy
has become much worse. Prior to my work , checking the %T accuracy with NIST SRM2031
90% nominal, showed error at 500nm to be about 0.5%T.
After my work the error is approx 2%T.
I have run wavelength cal and set scan gain numerous times.
Things I did before this showed up;
-replaced lamp
-aligned lamp
-minor smudges cleaned on filter wheel
-removed and cleaned sensor
-aligned #2 slit
-moved grating mirror beam up to center on slit at 550nm (screw adjust on mirror base)
I have run the performance tests, all look good.
Absorbance noise; 0.000027 avg SD, spec <=0.0002A
Wavelength Diff: 0
Repeatability diff; 0
Resolution 1.7390nm (spec<=1.8nm)
Baseline flatness, RMS 0.0004 (spec <=0.001A)
Stability at 340nm , 60 min. delta abs -0.00143 spec <0.003A
I tried undoing most of what I did, including;
-reinstalling the old halogen lamp
-moving the grating mirror beam back to half down off the #2 slit
-several setscan gain
-several power on/off
Why is this suddenly 4 times worse than before? Where do I look to fix?
Thanks!